MEA suggestion
Page 1 of 1
MEA suggestion
Hello All,
I need a suggestion as to what type of MEA I should order for a specific task. I am trying to do optical stimulation of mouse cortical neurons. I am tending towards an ITO MEA to reduce photoelectric effects due to stray laser light. I was wondering what kind of electrode spacing and pad diameter would be best? So far, I am tending towards 200 um spacing with 10 um diameter electrodes, but I am not sure if the high impedance caused by the smaller 10um pads will cause a low Signal/Noise ratio? Should I go with 30 um pads instead?
Thanks,
Steve
I need a suggestion as to what type of MEA I should order for a specific task. I am trying to do optical stimulation of mouse cortical neurons. I am tending towards an ITO MEA to reduce photoelectric effects due to stray laser light. I was wondering what kind of electrode spacing and pad diameter would be best? So far, I am tending towards 200 um spacing with 10 um diameter electrodes, but I am not sure if the high impedance caused by the smaller 10um pads will cause a low Signal/Noise ratio? Should I go with 30 um pads instead?
Thanks,
Steve
mcs- Posts : 518
Join date : 2008-06-10
Re: MEA suggestion
Dear Steve,
are you planning to do the optical stimulation from above ot from
below? ITO is a good idea in any case, but if you stimulate from below, I would suggest to use the thin ITO MEAs. Only problem is, there are only 100/10 or 200/30 available. However, the 30 or 10µm electrodes do not make much of a difference concerning signal to noise ratio. The 30µm are usually better for electrical stimulation.
Best regards Frank, MCS
>Hello All,
>
>I need a suggestion as to what type of MEA I should order
>for a specific task. I am trying to do optical stimulation
>of mouse cortical neurons. I am tending towards an ITO MEA
>to reduce photoelectric effects due to stray laser light. I
>was wondering what kind of electrode spacing and pad
>diameter would be best? So far, I am tending towards 200 um
>spacing with 10 um diameter electrodes, but I am not sure if
>the high impedance caused by the smaller 10um pads will
>cause a low Signal/Noise ratio? Should I go with 30 um pads
>instead?
>
>Thanks,
>
>Steve
are you planning to do the optical stimulation from above ot from
below? ITO is a good idea in any case, but if you stimulate from below, I would suggest to use the thin ITO MEAs. Only problem is, there are only 100/10 or 200/30 available. However, the 30 or 10µm electrodes do not make much of a difference concerning signal to noise ratio. The 30µm are usually better for electrical stimulation.
Best regards Frank, MCS
>Hello All,
>
>I need a suggestion as to what type of MEA I should order
>for a specific task. I am trying to do optical stimulation
>of mouse cortical neurons. I am tending towards an ITO MEA
>to reduce photoelectric effects due to stray laser light. I
>was wondering what kind of electrode spacing and pad
>diameter would be best? So far, I am tending towards 200 um
>spacing with 10 um diameter electrodes, but I am not sure if
>the high impedance caused by the smaller 10um pads will
>cause a low Signal/Noise ratio? Should I go with 30 um pads
>instead?
>
>Thanks,
>
>Steve
mcs- Posts : 518
Join date : 2008-06-10
Re: MEA suggestion
I think we will end up getting a mix of thin and standard ITO MEAs. Thank you for your suggestion.
mcs- Posts : 518
Join date : 2008-06-10
Similar topics
» Suggestion regarding MC_Rack
» a suggestion about the spike sorter
» A little suggestion for the release of MCSUSBNET api dll
» MC_Rack suggestion thread (please discuss)
» a suggestion about the spike sorter
» A little suggestion for the release of MCSUSBNET api dll
» MC_Rack suggestion thread (please discuss)
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum