Detection of electrical signal
Page 1 of 1
Detection of electrical signal
I am very interested in you multielectrode system. I understand the electrodes are made from indium tin oxide. This, I think, is a very basic question, but how is current injected into a sample, or how are local field potentials measured? When using a silver wire electrode I understand how the flow of chloride ions from solution form AgCl on a silver wire microelectrode, so how is the equivalent electrical signal detected with the indium tin oxide?
Roman
Roman
mcs- Posts : 518
Join date : 2008-06-10
Re: Detection of electrical signal
Dear Roman
The main difference is that you measure extracellular potentials with a passive metal conductor (= electrode). The extracellular space is conductive as well, and though the resistance is very low, it is not zero. According to Ohm's law (U=RxI, where U=voltage, R=resistance, and I=current), the extracellular current results in a small voltage that can be measured with extracellular electrodes.
The electrode material is Titanum Nitride (TiN). The main advantage of this material is that it has a high surface area (forming folded microstructures) and therefore has a low impedance and a high signal-to-noise ratio.
I hope this helps, Christine
The main difference is that you measure extracellular potentials with a passive metal conductor (= electrode). The extracellular space is conductive as well, and though the resistance is very low, it is not zero. According to Ohm's law (U=RxI, where U=voltage, R=resistance, and I=current), the extracellular current results in a small voltage that can be measured with extracellular electrodes.
The electrode material is Titanum Nitride (TiN). The main advantage of this material is that it has a high surface area (forming folded microstructures) and therefore has a low impedance and a high signal-to-noise ratio.
I hope this helps, Christine
mcs- Posts : 518
Join date : 2008-06-10
Re: Detection of electrical signal
Hi Christine,
I've been reading your reply with interest.
Can you confirm that MEAs are therefore suitable for recording cell membrane potential variations? I understand that this would imply an indirect measurement of the plasma membrane potential through a direct measure of the extracellular potential between the cell and the close electrodes. Of course I should keep the electrodes at a fixed voltage.
Am I right?
Thank you for your kind comments and suggestions!
Arianna
I've been reading your reply with interest.
Can you confirm that MEAs are therefore suitable for recording cell membrane potential variations? I understand that this would imply an indirect measurement of the plasma membrane potential through a direct measure of the extracellular potential between the cell and the close electrodes. Of course I should keep the electrodes at a fixed voltage.
Am I right?
Thank you for your kind comments and suggestions!
Arianna
mcs- Posts : 518
Join date : 2008-06-10
Similar topics
» Detection of electrical signal
» Realtime signal detection and feedback for in vivo and in vitro systems
» I am not getting a signal, help
» Spike duration/Field potential duration (FPD) calculation
» Losing signal
» Realtime signal detection and feedback for in vivo and in vitro systems
» I am not getting a signal, help
» Spike duration/Field potential duration (FPD) calculation
» Losing signal
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum
|
|