Can Mea120 system record data with MEA in incubator
Can we record data from the MEA while it is in the incubator?
If so what precautions should we observe to ensure that the electronics that mate to the MEA are protected?
Any suggestions would be appreciated.
Thanks
If so what precautions should we observe to ensure that the electronics that mate to the MEA are protected?
Any suggestions would be appreciated.
Thanks
dbumn- Posts : 2
Join date : 2009-05-22
Re: Can Mea120 system record data with MEA in incubator
Hello,
are we talking about a regular MEA amplifier(s), and an incubator with 100% humidity?
Regards Frank MCS
are we talking about a regular MEA amplifier(s), and an incubator with 100% humidity?
Regards Frank MCS
Frank MCS- Posts : 188
Join date : 2008-07-14
Can Mea120 system record data with MEA in incubator
Hello,
Yes we have the regular MEA amplifiers, and currently the incubator has 85% humidity. Is there a humidity that the amps could tolerate?
Yes we have the regular MEA amplifiers, and currently the incubator has 85% humidity. Is there a humidity that the amps could tolerate?
dbumn- Posts : 2
Join date : 2009-05-22
Re: Can Mea120 system record data with MEA in incubator
Hi,
in theory, if everything has the same temperature and no condensation happens, things could work fine, but I would not try that on an expensive piece of hardware. And obviously, any resulting damage would not be on guarantee. I would suggest using a dry incubator and cover the cells with the membranes that prevent evaporation (ALA-MEA-MEM, http://www.multichannelsystems.com/products-mea/product-details/products/189/ala-mea-mem5.html). That should work.
Frank
in theory, if everything has the same temperature and no condensation happens, things could work fine, but I would not try that on an expensive piece of hardware. And obviously, any resulting damage would not be on guarantee. I would suggest using a dry incubator and cover the cells with the membranes that prevent evaporation (ALA-MEA-MEM, http://www.multichannelsystems.com/products-mea/product-details/products/189/ala-mea-mem5.html). That should work.
Frank
Frank MCS- Posts : 188
Join date : 2008-07-14
Similar topics
» Read mcd data in a 64bit system by matlab
» 'Error reading raw data stream electrode raw data. Excluded from conversion'
» MEA system set-up
» ME-16-USB system
» Multiple STG1008 System
» 'Error reading raw data stream electrode raw data. Excluded from conversion'
» MEA system set-up
» ME-16-USB system
» Multiple STG1008 System
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum