data aquisition
Page 1 of 1
data aquisition
We are currently running two MEA systems and analysis the spike data in neuroexplore. When you import the files into neuro explorer you get a list of the channels which were active ie c12, c12 c13, c15, c15. If you are comparing multiple files ( we have been having to have multiple files since the computer keeps crashing during our recording see previous message) the second file (which could be a drug treatment) may import c12, c12, c13, c13, c15. when this occurs I have no sure way of knowing whether the c13 from the first file aligns with the first or second c13 in the second file. We have tried to replay and re-record the two meas independent of one another but the computer crashes during replaying.
Is it possible for MCRack to save the data from each mea as two seperate files? Or for us to import it as two seperate files?
thanks
sally
Is it possible for MCRack to save the data from each mea as two seperate files? Or for us to import it as two seperate files?
thanks
sally
mcs- Posts : 518
Join date : 2008-06-10
Re: data aquisition
Dear Sally
You cannot assign channels to two different data files with MC_Rack. But the experiment you describe should of course work without any computer crashes. So I would rather try to analyze the crashing problem than to find a workaround.
Best regards, Christine
You cannot assign channels to two different data files with MC_Rack. But the experiment you describe should of course work without any computer crashes. So I would rather try to analyze the crashing problem than to find a workaround.
Best regards, Christine
mcs- Posts : 518
Join date : 2008-06-10
Re: data aquisition
Dear Sally
Concerning the problem you have with the Neuroexlorer and the double MEA layout A and B: The thing is, that we informed the Neuroexplorer company about the changes in the MC_Rack data format. So far, they have not implemented the tags for MEA A and B. Maybe it would not be too bad if you contacted the Neuroexplorer company as well, to push things.
Best regards, Christine
Concerning the problem you have with the Neuroexlorer and the double MEA layout A and B: The thing is, that we informed the Neuroexplorer company about the changes in the MC_Rack data format. So far, they have not implemented the tags for MEA A and B. Maybe it would not be too bad if you contacted the Neuroexplorer company as well, to push things.
Best regards, Christine
mcs- Posts : 518
Join date : 2008-06-10
Similar topics
» How to analyse my data
» 'Error reading raw data stream electrode raw data. Excluded from conversion'
» How to analyze the data from MEA
» How to analyze the data from MEA
» get data from analog 2
» 'Error reading raw data stream electrode raw data. Excluded from conversion'
» How to analyze the data from MEA
» How to analyze the data from MEA
» get data from analog 2
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum