Spike detection error
Page 1 of 1
Spike detection error
I frequently encounter the following error message when using spike detection, specially during stimulations.
"System is at its performance limit (due to spike detector) and cannot process all data. Try to do 1 or more of the following: (i) remove unused displays, (ii) make displays smaller and view only 1-6 channels, (iii) remove analysis instruments and do off-line analysis. Please watch the status bar for information on system load."
This is seen irrespective of the number of displays that i have opened or whether the spike detection is done during recording or post-recording. The status bar at such times, always shows spikes (MB) as 9/8 (the numerator is >8,) and it stops proceeding further. Sometimes, the program exits by itself. As a result, I am not able to get a full minute recording every time I stimulate. Is there any way to remedy this? Even changing the thresholds does not help. When stimulation is not done, this error is not seen much and not a problem. Probably noise and the stimulus artifacts interfere and exceed the spike detector's capacity, but I hope there is some solution to it.
Thank you,
Sincerely,
K Fozdar.
"System is at its performance limit (due to spike detector) and cannot process all data. Try to do 1 or more of the following: (i) remove unused displays, (ii) make displays smaller and view only 1-6 channels, (iii) remove analysis instruments and do off-line analysis. Please watch the status bar for information on system load."
This is seen irrespective of the number of displays that i have opened or whether the spike detection is done during recording or post-recording. The status bar at such times, always shows spikes (MB) as 9/8 (the numerator is >8,) and it stops proceeding further. Sometimes, the program exits by itself. As a result, I am not able to get a full minute recording every time I stimulate. Is there any way to remedy this? Even changing the thresholds does not help. When stimulation is not done, this error is not seen much and not a problem. Probably noise and the stimulus artifacts interfere and exceed the spike detector's capacity, but I hope there is some solution to it.
Thank you,
Sincerely,
K Fozdar.
mcs- Posts : 518
Join date : 2008-06-10
Re: Spike detection error
Dear K. Fozdar
it looks like that the spike detector level is within the noise level and thus leading to the overload.
Please check the "Spike detector thresold" in the display window when you are looking at spikes and make sure that the spike detection level is OUSIDE the general noise level.
best regards
Michael Fejtl
Dr.Michael Fejtl, CSO
Multi Channel Systems GmbH
it looks like that the spike detector level is within the noise level and thus leading to the overload.
Please check the "Spike detector thresold" in the display window when you are looking at spikes and make sure that the spike detection level is OUSIDE the general noise level.
best regards
Michael Fejtl
Dr.Michael Fejtl, CSO
Multi Channel Systems GmbH
mcs- Posts : 518
Join date : 2008-06-10
Similar topics
» Spike detection- Threshold
» FIND toolbox for spike detection in MATLAB
» Why no drop down selection for units for Trigger Detection Level?
» Manually adjusting spike detection threshold in MC_Rack with mouse
» Spike duration/Field potential duration (FPD) calculation
» FIND toolbox for spike detection in MATLAB
» Why no drop down selection for units for Trigger Detection Level?
» Manually adjusting spike detection threshold in MC_Rack with mouse
» Spike duration/Field potential duration (FPD) calculation
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum
|
|