MC RACK stops recording - Electrode Raw data in watchog
Hi all,
Recently during our recordings sessions, MC Rack stops and prints the message "Limit on System performance due to electrode raw data in watchdog". I have tried not using any display, restarting the computer and closing all programs but MC Rack but the problem continues. Any help would be appreciated
Recently during our recordings sessions, MC Rack stops and prints the message "Limit on System performance due to electrode raw data in watchdog". I have tried not using any display, restarting the computer and closing all programs but MC Rack but the problem continues. Any help would be appreciated
perez087- Posts : 1
Join date : 2009-09-24
MC Rack Problems...
Hi,
to investigate what is going wrong there we would need to know what PC you are using, how many channels you are recording, sampling frequency... and we would need to see your rach file to try to reproduce the problem.
Please mail it to support@multichannelsystems.com and we will take care of it.
Greetings
Thomas
to investigate what is going wrong there we would need to know what PC you are using, how many channels you are recording, sampling frequency... and we would need to see your rach file to try to reproduce the problem.
Please mail it to support@multichannelsystems.com and we will take care of it.
Greetings
Thomas
ThomasMCS- Posts : 71
Join date : 2008-07-16
Similar topics
» 'Error reading raw data stream electrode raw data. Excluded from conversion'
» Problem with MC Rack data Replayer
» recording bipolar hook wire electrode
» MC Rack software crashing with big files
» Trigger an external stimulator using MC Rack
» Problem with MC Rack data Replayer
» recording bipolar hook wire electrode
» MC Rack software crashing with big files
» Trigger an external stimulator using MC Rack
Page 1 of 1
Permissions in this forum:
You cannot reply to topics in this forum